[Sneap] DILO (MACHINE FOR DRYING SF6) PROBLEM
Anke Rieck
arieck at leibniz.uni-kiel.de
Tue Jan 23 11:40:10 EST 2007
Dear Riyad,
There's a contact pressure gauge behind the compressor that protects it
from a too high output pressure. It can be adjusted, in our case it's
set to 40 bars. With a closed valve the 40 bars are reached within seconds.
Regards
Anke
Anke Rieck
Leibniz-Laboratory for Radiocarbon Dating and Isotope Research
University Kiel, Germany
Tel +49 (0)431 880 7375
> ------------------------------------------------------------------------
>
> From: sneap-bounces at tunl.duke.edu [mailto:sneap-bounces at tunl.duke.edu]
> On Behalf Of riyad sha
> Sent: Saturday, January 20, 2007 2:17 AM
> To: sneap at daytona.tunl.duke.edu
> Subject: [Sneap] DILO (MACHINE FOR DRYING SF6) PROBLEM
>
>
>
> Hi sneapers
>
> I tried to dry SF6 gas of our accelerator machine during the weekend ,
>
> When I returned, I found the DILO machine automatically switched off
> may be because the overheating resulting from the long operation .
>
> When I tried to apply the procedure again and again, the machine
> have been worked just for a few seconds every time and switched off .
>
> In reference to the DILO manual I didnt find any thing a bout this
> case .
>
> By testing the control power there arent any problem with the control
> power ,may be there is a problem with temperature sensors which is
> compact with the compressors .
>
> Pleas colleagues any one encounter this problem or have the experience
> a bout that .
>
> Best regards.
>
>
>
> Riyad Shaban
> Electronic engineer in AECS,
> Accelerators Department
> Address: Damascus-Syria-P.O.B:12711
> TEL :( +963) (11) (3921506)
> MOBILE :( 963) (93956292)
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://mail.tunl.duke.edu/pipermail/sneap/attachments/20070123/c101595d/attachment.htm
More information about the Sneap
mailing list