You want what to be portable? The transformer itself, the power supply and transformer, the entire deck? None of these are really specified.

Looking at your request, I think you are asking for a 120VAC primary, some lower secondary VAC, and 50kVDC isolation primary to secondary. You plan to make a low voltage DC power supply of order 1W using the isolated secondary. The deck is at some DC elevation of order 50kV.

This is not too tough to make if you can't buy it. If there is a local transformer shop contact them about what is required to make the transformer without isolation. Ask for a rectangular core and circular windings. The windings can be made of small gauge wire because the power requirement is so small. Then tell them the core you want has to have legs long enough to provide 5" of track length on each side of the winding to the top and bottom of the core. That's 10kV per inch track length on polyethylene sheet. You also need to stuff at least .050" of polyethylene between the primary and secondary so the gap between the windings is important. That's 1kV per mil dielectric strength. A wrap of .050" of polyethylene around the secondary ensures that the secondary doesn't arc to the opposite leg. In short, the secondary must look along 10kV per inch before it sees the top and bottom of the core. 
It must also look through 1kV per mil of polyethylene before it sees the primary or the opposite leg of the core.

The core can be made of Tempel 29 gauge (0.014") sheared sheets of silicon steel or equivalent. Stagger the strips at the corners to overlap completely. I make core clamps out of aluminum angle. Remember, no conductors other than the windings encircle the core.

Tempel saturates at about 1.6 Tesla so designing it to operate at 1.0 Tesla is way safe. If you want a 1" cross section, that is 6.4e-4m^2. Phi max is area times B max or 6.4e^-4 times 1 tesla or 6.4e^-4 webers. Omega is at 60Hz so dphi/dt=omega phi max or 2.4 volts peak per turn which is 0.17 Vrms per turn. For 120 VAC this is 700 turns for the primary. A small gauge wire will constitute some significant resistance. The secondary turns are determined by the desired voltage.

The exciting current depends on the magnetic field path in the core. This is a long core and it will consequently take some exciting current. If the legs are 12" long and the top and bottom are 4" long the path length is 32" or 0.8m.
The exciting current ie=BL/uNp where u for Tempel is 0.004. Then ie is 0.28Ap.
Play around with the dimensions and wire gauges until you are satisfied.

-Greg

Gregory C. Harper               Phone:(206) 543-4084 (Office)
Research Engineer/Scientist IV        (206) 543-4062 (Lab)      
University of Washington        FAX:  (206) 685-4634
CENPA NPL Bldg.
Box 354290                      email: gharper at u.washington.edu
Seattle, WA 98195-4290

On Tue, 18 Sep 2007, Chris Westerfeldt wrote:

> The ATF-SNEAP 2006 meeting was held in Australia and information is available at:
> http://www.ansto.gov.au/nugeo/conference/private/ATF_SNEAP.htm
>
> ++++++++++++++++++++++++++++++++++++
>
> The SNEAP 2007 meeting will be hosted by Brigham Young University in Salt Lake City Utah. The meeting web site is located at:
> http://webpub.byu.net/wsd/.
> The 41st SNEAP Conference
> Brigham Young University
> October 7 - 11, 2007
>
> Please note:
> DUE TO THE ORIGINAL ANNOUNCEMENT BEING LOST IN CYBER-SP[ACE, THE
> SUBMISSION DATE HAS BEEN EXTENDED TO September 15th
>
> Call for Papers:
>
> Brigham Young University in sponsoring the 41st Conference of the
> Symposium of North Eastern Accelerator Personnel (SNEAP) on October 7-
> 11, 2007. The conference will be held at the Embassy Suites Hotel in
> Salt Lake City, Utah.
>
> The SNEAP Conference is directed toward scientists and technical staff
> who are involved in developing, upgrading, and operating electrostatic
> accelerators. Further information about the conference is available at
> the conference website:
>
>
> http://webpub.byu.net/wsd.
>
>
> We invite presentations on topics relating to accelerator design,
> operation, and improvement, and to unresolved problems with accelerator
> systems. Please send a title and abstract of your proposed presentations
> to sneap at byu.edu by September 15.
>
> Contacts:
> Scott Daniel       wsd at byu.edu
> Lawrence Rees      sneap at byu.edu
>




More information about the Sneap mailing list