Stuart,
    Carbon film resistors do age in high voltage operation ­ the resistance
declining with time.  Finding the value
at 50% is very unusual and I suspect some damage from sparks.  I have
attached a drawing of our KN focus power supply
circuit made in 1997.  The components we used are identified on the drawing.
I also attach a sketch of the Focus output protection.
This helps prevent damage to the supply from tank sparks.
    Regarding the Palladium leak ­ yes we have gone through a number of them
over the years.  Sometimes they cold flow ­ which is more of a nuisance but
typically they will fail to pass hydrogen and have to be run hotter and
hotter until they no longer work at all.  There is a procedure that I have
seen for 
cleaning them and we have actually replaced the Palladium thimbles in the
past.  The last time I talked to HVEC about this ­ maybe ten or fifteen
years ago, I was told that they could no longer obtain quality thimbles and
were no longer making these leaks.  We switched to the TM leaks in our JN
and KN.  We tried the Palladium leaks made by Potentials Inc. but never got
satisfactory performance from them.
I  hope this is useful.
Regards,
- Chris
-- 

Chris R. Westerfeldt
Research Scientist / T.U.N.L. Radiation Safety Manager
Duke University Physics Department &
Triangle Universities Nuclear Laboratory
Science Drive, Box 90308
Durham, NC  27708-0308
Tel: (919) 660-2600
Fax: (919) 660-2634
Email: Cwest at Tunl.Duke.Edu



From: "Van Deusen, Stuart B" <sbvande at sandia.gov>
Reply-To: Symposium of Northeastern Accelerator Personnel
<sneap at tunl.duke.edu>
Date: Fri, 28 Sep 2007 09:08:25 -0600
To: <SNEAP at TUNL.DUKE.EDU>
Conversation: Focus resistor, Pd leak on AN2500
Subject: [Sneap] Focus resistor, Pd leak on AN2500

The ATF-SNEAP 2006 meeting was held in Australia and information is
available at:
http://www.ansto.gov.au/nugeo/conference/private/ATF_SNEAP.htm

++++++++++++++++++++++++++++++++++++

The SNEAP 2007 meeting will be hosted by Brigham Young University in Salt
Lake City Utah. The meeting web site is located at:
http://webpub.byu.net/wsd/.
The 41st SNEAP Conference
Brigham Young University
October 7 - 11, 2007

Please note:
DUE TO THE ORIGINAL ANNOUNCEMENT BEING LOST IN CYBER-SP[ACE, THE
SUBMISSION DATE HAS BEEN EXTENDED TO September 15th
 
Call for Papers: 

Brigham Young University in sponsoring the 41st Conference of the
Symposium of North Eastern Accelerator Personnel (SNEAP) on October 7-
11, 2007. The conference will be held at the Embassy Suites Hotel in
Salt Lake City, Utah.

The SNEAP Conference is directed toward scientists and technical staff
who are involved in developing, upgrading, and operating electrostatic
accelerators. Further information about the conference is available at
the conference website:
 

http://webpub.byu.net/wsd.


We invite presentations on topics relating to accelerator design,
operation, and improvement, and to unresolved problems with accelerator
systems. Please send a title and abstract of your proposed presentations
to sneap at byu.edu by September 15.
 
Contacts:
Scott Daniel       wsd at byu.edu
Lawrence Rees      sneap at byu.edu

Hello all 

The output from the focus power supply on an AN2500 goes through a resistor
on its way to being clipped to the focus plane on the column.  The HVEC
circuit diagram shows this as a 2 Mohm resistor.  When I started operating
my machine 11 years ago, it was equipped with a string of 2W carbon
composition resistors totaling about 11Mohm.  I recently managed to blow
these resistors up (probably by arcing over) by running the focus voltage
too high while the machine was in normal operation.  I replaced the string
of carbon composition resistors with a single thick carbon film resistor,
rated at 5W and 20 KV.  I find that after some use, the value of the
resistor drops, in the most recent case by almost half.  I don't really
understand what is happening to the resistor, and if it will ultimately fail
open or not.

My question is, what do you all put in when you replace this resistor?

A second question concerns the palladium leak used for hydrogen in the
AN2500 (and I'm sure other machines as well).  Does anyone have experience
with these leaks failing, and if so, what is the failure mode?  On both this
leak and the thermo mechanical leak used for He, what is the expected leak
rate when a good leak is fully "closed"?

Thank you for your help.

Stuart Van Deusen 


-----------------------------------------------------------
Stuart B. Van Deusen
Radiation Solid Interactions and Processing
Dept. 1111, M/S 1056
Sandia National Laboratories
P.O. Box 5800 
Albuquerque, NM  87185-1056
USA 
PH (505)844-7782 
FAX (505)844-7775 
e-mail: sbvande at sandia.gov

address for package delivery:

US NNSA c/o Sandia Natl. Labs
ATTN: Stuart Van Deusen
Bldg. 884, Rm. 11 
1515 Eubank S.E. 
Albuquerque, NM  87123
USA 


_______________________________________________
Sneap mailing list
Sneap at daytona.tunl.duke.edu
http://daytona.tunl.duke.edu/mailman/listinfo/sneap

-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://daytona.tunl.duke.edu/pipermail/sneap/attachments/20070928/9d1c42df/attachment-0001.htm 
-------------- next part --------------
A non-text attachment was scrubbed...
Name: KNFocus.pdf
Type: application/octet-stream
Size: 36517 bytes
Desc: not available
Url : http://daytona.tunl.duke.edu/pipermail/sneap/attachments/20070928/9d1c42df/attachment-0002.obj 
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Focusprot.pdf
Type: application/octet-stream
Size: 29818 bytes
Desc: not available
Url : http://daytona.tunl.duke.edu/pipermail/sneap/attachments/20070928/9d1c42df/attachment-0003.obj 


More information about the Sneap mailing list