In response to the question regarding the amount of water cooling needed for the
EN accelerator cooling ( Belt charging system ).
Here at Kansas State University before the installation of the our pelletron system the water pressure
at the entrance of the tank was approx. 70 PSI and the chilled water temperature was 55 to 60
degrees F. I cannot tell you what the actual water flow was through the cooling coil inside the
tandem due to the return pressure, etc. I can tell you that our normal return pressure in our chiller
system is 15 PSI.
Bob Krause
James R. Macdonald lab.
Kansas State University
************************************************************************************