Remove the two side plates (screws).  Remove the sample vial (not the empty 
reference vial).by pushing it through.  .  We can't remember for sure, but 
believe that the fragile coil was very tight on the original vial.  If 
necessary, you could try refilling the vial in place.  We have sometimes 
broken the vials.  Fill the vial with a millimolar solution of MnCl4.4H20 
(try 1.1 grams per 100 ml water).  We now seal the vials with epoxy or 
Q-dope rather than the original stopper.  Insert the vial and replace the 
metal plates.  Voila!

    Art Haberl

----- Original Message ----- 
From: "Guy Garty" <gyg2101 at columbia.edu>
To: <sneap at tunl.duke.edu>
Sent: Friday, March 07, 2008 11:53 AM
Subject: [Sneap] Spectromagnetics NMR gaussmeter


> Hi all,
> We have an old Spectromagnetics NMR gaussmeter (model 5002). All but one
> of its probes have died on us. We disassembled one of the dead probes and
> it does not seem obvious how to fix it, as all electrical contacts seem to
> be intact. We suspect that the "sample solution of Proton ion" (as it is
> described in the manual) has evaporated or leaked out.
>
> Does anybody know how to fix these probes or where to get replacements?
> The probes themselves have no markings, but are connected to the
> electronics by 3 BNC connectors (one of which has 2 pins) and a 5 pin 
> round
> Amphenol connector.
>
> Thanks,
> Guy Garty
> _________________________________________________________
> RARAF, Nevis Laboratory       Phone: (914) 591-9244
> Columbia University           Fax: (914) 591-9405
> 136 S. Broadway               Email: gyg2101 at columbia.edu
> Irvington, NY 10533           Web Page: http://www.raraf.org
>
> _______________________________________________
> Sneap mailing list
> Sneap at daytona.tunl.duke.edu
> http://daytona.tunl.duke.edu/mailman/listinfo/sneap 



More information about the Sneap mailing list