[Sneap] NEC pelletron C stripper foil question
Albert Zondervan
A.Zondervan at gns.cri.nz
Wed Apr 29 18:16:01 EDT 2009
Lucas
Being most familiar with Accelerator Mass Spectrometry I can share some
information that might help. In AMS, one quintessential parameter is the
efficiency of counting ions of the rare isotope of interest. The
acceleration step, which is needed to destroy ion beams that could
interfer with that counting, involves one component of the overall
efficiency: the percentage of negative ions that get converted to the
positive charge state that is selected for further
momentum/velocity/energy-analysis and ultimately for counting with a
particle detector. For AMS, one selects and optimises terminal voltage to
get the highest possible charge-state fraction. (The finer points of
optimisation do not need mentioning here.) That percentage is a function
of projectile velocity, its nuclear charge, stripper medium, and stripper
thickness. The process, due to the stochastic nature of ions travelling
thru the medium thereby having numerous charge-exchange interactions with
it, can be described by charge-state distributions. These be found in the
literature and websites etcetera for the equilibrium situation.
Equilibrium means here: Adding more thickness does not change the
distribution. Of course thickness has a strong effect on scattering
(energy, angular) and thus influences beam optical transport.
Most AMS systems have recirculating-gas strippers. Some older systems,
like my own, still have C stripper foils. Quite a few systems out there
probably have both. The important message others can comment on much
better is that there are applications where it pays to work with foils
rather than with gas. But I don't think getting the highest possible
energy for C ions using a 3 MV tandem is helped by using foils rather than
gas. I think if you'd experiment with the nitrogen gas thickness at fixed
terminal voltage, you will see that there is an optimum thickness
(pressure) at which the beam intensity is maximum. This probably is
charge-state dependent. Certainly it will depend on how well beam
transport is taken care of. From that "angle" you can probably argue that
replacing the gas by foil is not going to make a major difference.
Pointer: [Shima et al. "Equilibrium charge fractions" Atomic Data and
Nuclear Data Tables, Vol 51, Number 2, July 1992, page 189]. It shows the
fraction (%) of C projectiles thru C foil, as function of velocity-squared
(MeV/u) and for the different charge states (e). To generate the
bottom-end tail percentage of 1%, you need 0.2 MeV/u which converts to 2.4
MeV C-12 ions at the terminal. At 3 MeV, F(q=5+) = 3% and F(q=6+) << 1%.
So, at first glance you might be able to have some 5+ charge state at 3
MV, even with gas. Careful momentum analysis with the HE-side magnet will
(dis)prove this. And then the question is if the overall low efficiency
for this charge state found at your highest possible terminal voltage can
be compensated by injecting with a stiffer beam, as the Pelletron is
limited by how much beam can be injected (charge can be dumped at the
terminal per unit time). If you need for your IBA applications only a few
nA, then you might get away with injecting one or several uA... Ask NEC
where the limits are for your Pelletron-system.
Upshot:
1/ A foil stripper might help, but probably not give you a substantial
gain.
2/ You might be able to get a nA of carbon in the 5+ charge state.
3/ Your terminal voltage is way too low to reach the 6+.
Good luck with the search: both in literature and by experimentation !
cheers
Albert
Lucas Phinney <lucas.phinney at gmail.com>
Sent by: sneap-bounces at tunl.duke.edu
28/04/2009 04:21 a.m.
Please respond to Symposium of Northeastern Accelerator Personnel
To: SNEAP <sneap at tunl.duke.edu>
cc:
Subject: [Sneap] NEC pelletron C stripper foil question
Hello,
I am a graduate student at the University of North Texas working in the
Ion Beam Modification and Analysis Laboratory (IBMAL) with Dr. Jerry
Duggan. I am trying to get higher charge states of a C beam (+5 and +6),
with a 3 MV NEC 9SDH-2 Pelletron. We are currently using the Nitrogen gas
stripper and I can only get up to +4 charge state, and I was hoping that
by going to the solid foil I would be able to get the higher charge
state. I was wondering if anyone had experience using C foil strippers in
the tandem and could give me some pointers. I was specifically wondering
what thickness would be best, but any information would be appreciated.
Thanks
--
Lucas Phinney
Ion Beam Modification and Analysis Laboratory
Department of Physics
University of North Texas
(940) 565-3336_______________________________________________
Sneap mailing list
Sneap at daytona.tunl.duke.edu
http://daytona.tunl.duke.edu/mailman/listinfo/sneap
Notice: This email and any attachments are confidential. If received in error please destroy and immediately notify us. Do not copy or disclose the contents.
More information about the Sneap
mailing list